On the other hand, your separation at 0.4 mL/min on the 4.6 x 150 mm, 5 um column would translate directly to a flow rate of 0.083 mL/min on the 2.1 x 50 mm column. If you raised that flow rate to 0.15 mL/min to get to higher linear velocity of 1.4 mm/sec, your backpressure would be more like 235 bar at 25C. With the minimized extracolumn volume and at that flow rate you ought to get much better results.
With your current system, you will need the 2 uL flow cell for DAD or the 1 uL micro flow cell for VWD along with the 0.12 x 180 mm capillaries from ALS to column and column to flow cell. Changing the needle seat capillary to 0.12 mm ID will help a little too.
If you can eliminate the connection of the flow cell to the union, and connect the tubing from the flow cell directly to the column outlet, you'll fare better. If solubility permits, dilute your sample with water or aqueous you're using by 1:1, 1:2 or 1:3 and inject the respective larger injection volumes to account for the dilution. That can also have a significant effect on plates. If you can make the tubing from the flow cell to the mass spec be 75-100 um ID PEEK fused silica (SGE, Upchurch/IDEX) that would also help narrow your peaks in the MS chromatogram.
I hope this helps!