Blossom » Fri Feb 13, 2009 4:05 pm
													
 
						
					 
					
						If you want to use a 0.32mm id column on a mass spec you need to make sure that ms pumping system can handle the flow rate you want to use. If the flow rate is too high for the MS then you must split the flow between the mass spec and the column so that the flow to the ms is reduced to an acceptable level. 
The ms is at vacuum, inlet pressure controllers (electronic or otherwise) on GCs are not designed to maintain pressures below ambient pressure  and are not very good at maintaining pressure near their lower limits. A 30m x 0.32mm column does not have sufficient resistance to flow to ensure that at optimum flow rates the inlet pressure is above ambient, therefore to use a 0.32mm column you need to increase its resistance by either making it longer or by adding on a length of inert column of a smaller diameter. You add this piece of column not at the beginning but at the end. Calculate the diameter and length of column required to give you about 1 atmosphere at the start of this piece of column when the other end is under vacuum. The result of this will be that, the inlet pressure required will be above ambient and in the range that the pressure controller can operate accurately. There will be no vacuum in the analytical column thus improving performance. It will improve the quality of splitless injection because the higher inlet pressure will reduce the volume of the vaporised sample. Lastly when you change columns all you need to do is lower the MS temp to about 50C, you won't need to turn of the pumps or vent it ( if you're quick!).
I've used this method (when I was young) for measuring ethanol using static headspace, an FID and a 0.53mm id column.
My advice is, throw the 0.32mm column away and get the smallest diameter column that has the capacity for your analytes, try 0.25 or 0.2. Good luck