Progress and responses

The phenol suite I'm looking at is phenol, mono and dimethyl phenols, nitrophenols, all the chlorophenols, nitrophenols and a couple of selected alkyl phenols.
Its a GC-MS method using a standard Agilent 6890/5973.

Derivitisation works well for the standards made from the same solvents used in the extraction so I am confident that they are moisure free. Unfortunately we don't have any KF equipment on site for any further tests.

Using additional derivitising agent worked, but required equal amounts of extract/BSTFA to work.

As things stand my manager has put a stop on further development for this project. He isn't happy using 1:1 of the derivitising agent so it looks like we will be going for using increased sample volumes and accepting the X10 dilution post elution as a necessary step and the increased LODs that will come from it.
I have one more idea left to try which is the aquasep cartridge from varian.

Sassman,
I like your idea of using diethylether and solvent changing. I'll probably come in this weekend and give that a shot.

As always, thank you all for your input. I'll post conclusions once the validation is complete.

Rich