XL » Thu Mar 26, 2009 6:44 am
HPLCSlave,
We finally got time to do your application. First tried was an aqueous compatible RP column. We had similar observation to yours - acetaldehyde and lactic acid co-elute. Changing mobile phase pH or organic solvent composition didn't improve the situation. Then we tried Acclaim Mixed-Mode WAX-1 (a reversed-phase and weak anion-exchange mixed-mode column). And it worked. Here are the conditions:
Column: Acclaim Mixed-Mode WAX-1, 5-micron, 4.6x150-mm
Mobile phase: 39.1 g acetonitrile, 950 g water, 2.30 g NH4H2PO4 (20 mmol), 0.660 g (NH4)2HPO4 (5.0 mmol)
Flow rate: 1 mL/min
Detection: UV at 210 nm
Result:
1. Acetaldehyde tR= 3.7 min (k' ~ 0.8
2. Lactic acid tR=4.2 min (k' ~ 1)
3. Acetic acid (oxidation impurity of acetaldehyde) tR=4.4 min (k'~1.1)
The use of a 4.6x250-mm column will further improve the resolution.
I hope this will help.