Damn it, you have triggered my couriosity
I just might have to spend some time on this at my lab to try this out. Most likely I will use a 6890+ GC with a 7694 HS connected to a volatile inlet and silcosteel parts in HS ie pretty much like yours. Our new HS ie same as yours are not all installed yet so acess is limited to those.
viewtopic.php?t=9658 this thread is not the same thing excatly but is related to that you sometimes have to struggle a bit to get that low level sensitivity with HS- GC.
So purely teorectical you should be able to detect benzene at those levels.Would not say it is just the FID. More the headspace technique itself and the method in use in combination with the low levels your looking at.
A total of 0,1ug in the vial is hard to analyse but a rough approximation could be: 50% of 0,1ug acrylonitrile ends up in 10ml gas phase, 1ml is injected with a split ratio of 1:5=> 1 ng (1000pg) on the column. As a comparision you typically can detect about 20pg toluen (that ends up on capillary column) with a FID. Sure there is a difference in the responsfaktor between the two compunds but roughly.
Unsure when I can squeez in this testing but I will let you know when/if it works.
One thing to recall is also that with increased % of water in DMF you "generally" increase the sensitivity, have no good reference on that apart from graphs provided by "old" HP.