Posted: Sat Dec 06, 2008 2:16 pm
Have you changed the injector frit??? Once it gets contaminated it is impossible to un-contaminate.
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://www.chromforum.org/
With the additional information, eg appearance after injection of a blank, new columns, 4 HPLC systems, problem since development, etc., it seems as if your problem may be contamination of the mobile phase. In that case, to confirm, I would suggest performing the test outlined by mbicking above. If it show the mobile phase components/system is the problem, you need to try alternative suppliers of chemicals, and flush all parts of your solvent system.The peaks that are coming off in the Blank are a carbon copy of the chromatography of the sample injection. We know this because we have injected(individual related subsatnces and active peak) them at high concentrations individually. The blank peaks (carryover/ghostpeaks) are at a much smaller level (AU 0.000-0.006) but as i said before nearly a carbon copy as if injected a sample except they are alot smaller.
I have seen some variability with our HPLC water too. We are doing ion chromatography in our lab, so I check the pH of the water before making the mobile, and it does vary. I'm guessing it is disolved gasses because we have an inline degasser and the variation doesn't seem to effect the Chromatogram. This might be elementary, but my second guess is tech error, ie: The water jug becoming contaminated with stock solution when making the mobile.Just for kicks have you tried water from a different source? I've come across times when there has been peaks from my milli-Q water and not from bottled water and vise versa.