I use these six PAH internal standards as surrogate. I got six surrogates recovery increasingly high in some soil samples. For example, the recovery of the first three surrogate were around 80-90%, the forth one went up to 120%, the fifth one were 150%, and eventually, the sixth one went up to 200%, 300% sometimes. We don't do PAH soil and water sample cleanup. We use SIM mode. These surrogate peaks looks in good shape.
I suppose it's because of matrix interference, because most time these problem happened with dark sample(few times but still had sometimes happened with clear sample). But, how come interference could have the same RT and same ions with these deuterated compounds?
Can someone give some idea?
Are you using internal standards at all? That may make a big difference in your results. Internal standard techniques tend to compensate for changes in concentration due to evaporation, fluctuations in instrument sensitivity, etc. My problem is that my internal standard recoveries fluctuate independently of target compounds, which is not what you are looking for in an internal standard.
Your problem may be due to precipitation of the internal standards in your vial prior to cracking or to degradation of these compounds in your calibration standards. If you compare a calibration standard you injected yesterday to an injection of that same standard when it was first prepared, is there a large difference in abundance?
It may just be a matrix interference problem. Some of the heavier compounds may be more difficult to detect in a dirty matrix or may fluctuate in recovery as you are seeing.
I highly recomment you start using internal standard calibration techniques if you are not currently doing so. Use the standard 8270 base neutral and phenol surrogates (nitrobenzene-d5, 2-fluorobiphenyl, p-terphenyl-d-14, 2-fluorophenol, phenol-d6, 2,4,6-tribromophenol) and use these six compounds as internal standards.