Page 2 of 2
Re:
Posted: Mon Jul 16, 2012 7:34 am
by amirtamboli
Hi Mattias
Below is an application for Tween 20 in IgG aqueous solution
using Cadenza HS-C18:
http://www.silvertonesciences.com/files/TI384E.pdf
In the chromatogram - Tween 20 detergent is retained while the
IgG antibody is excluded.
Perhaps it is possible to follow similar experimental condtions
and collect the 50kDa protein at the void volume (free of Tween 20).
Dear Mr Evans
Can you send me the above application, mentioned in link as it is not opening and i am in need of tween 20 application
Re: Tween 20 peak in SEC
Posted: Mon Jul 16, 2012 9:17 pm
by Vlad Orlovsky
Brian is no longer with the company as far as I know.
Posted: Fri Aug 03, 2012 4:44 am
by amirtamboli
can you please solve my problem
Posted: Fri Aug 03, 2012 5:36 am
by prepcolumn
amirtamboli
here is the application
http://www.imtaktusa.com/site_media/fil ... TI384E.pdf
Posted: Fri Aug 03, 2012 5:44 am
by amirtamboli
amirtamboli
suggested one already tried with different column and shown good result but it has tailing
Posted: Mon Aug 06, 2012 10:48 am
by biochemist
Hi Amir,
You could try incorporating 25% IPA in your mobile phase, then using fluorescence detection.
Development of a sensitive size exclusion HPLC method with fluorescence detection for the quantitation of recombinant human erythropoietin (r-HuEPO) aggregates
http://www.sciencedirect.com/science/ar ... 8506004092
The method is for Polysorbate 80, but do let us know if it works for Polysorbate 20 as well.
Posted: Mon Aug 06, 2012 11:19 am
by Mattias
Being the thread starter, maybe I should tell how it went...
We gave up and introduced a placebo solution in the method. So now we subtact the peak area of the tween peak in the placebo sample from the aggregates peak area in the sample.
Not the best solution, but we ran out of time. It is never too late to give up.
Posted: Tue Aug 07, 2012 1:28 am
by biochemist
Mattias,
Congratulations on solving your problem! Did you face any problems during the validation?
I'm curious - does the Tween 20 elute in a consistent way (size, shape and retention time)? We tended to get wildly different results for a placebo, depending on the Tween 20 concentration - sometimes it would be well-behaved, and sometimes not. What column are you using?
Posted: Tue Aug 07, 2012 6:16 am
by Mattias
The Tween 20 peak is fairly consistent in placebo samples, but the precision is pretty bad anyhow. The Tween peak is about 50% of the aggregate peak area in stability samples and close to 100% of the peak area in newly produced batches.
The validation was more focused on using the method as a limit test for aggregates, rather than calculating the precision at different concentrations. It was acceptable, but not fantastic. The column is a TSK column.