Tween 20 peak in SEC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

24 posts Page 2 of 2

Re:

Hi Mattias

Below is an application for Tween 20 in IgG aqueous solution
using Cadenza HS-C18:

http://www.silvertonesciences.com/files/TI384E.pdf

In the chromatogram - Tween 20 detergent is retained while the
IgG antibody is excluded.

Perhaps it is possible to follow similar experimental condtions
and collect the 50kDa protein at the void volume (free of Tween 20).
Dear Mr Evans
Can you send me the above application, mentioned in link as it is not opening and i am in need of tween 20 application
Brian is no longer with the company as far as I know.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
can you please solve my problem
can you please solve my problem

amirtamboli

here is the application

http://www.imtaktusa.com/site_media/fil ... TI384E.pdf
can you please solve my problem

amirtamboli

here is the application

http://www.imtaktusa.com/site_media/fil ... TI384E.pdf
suggested one already tried with different column and shown good result but it has tailing
Hi Amir,

You could try incorporating 25% IPA in your mobile phase, then using fluorescence detection.

Development of a sensitive size exclusion HPLC method with fluorescence detection for the quantitation of recombinant human erythropoietin (r-HuEPO) aggregates
http://www.sciencedirect.com/science/ar ... 8506004092

The method is for Polysorbate 80, but do let us know if it works for Polysorbate 20 as well.
Being the thread starter, maybe I should tell how it went...

We gave up and introduced a placebo solution in the method. So now we subtact the peak area of the tween peak in the placebo sample from the aggregates peak area in the sample.

Not the best solution, but we ran out of time. It is never too late to give up.
Mattias,

Congratulations on solving your problem! Did you face any problems during the validation?

I'm curious - does the Tween 20 elute in a consistent way (size, shape and retention time)? We tended to get wildly different results for a placebo, depending on the Tween 20 concentration - sometimes it would be well-behaved, and sometimes not. What column are you using?
The Tween 20 peak is fairly consistent in placebo samples, but the precision is pretty bad anyhow. The Tween peak is about 50% of the aggregate peak area in stability samples and close to 100% of the peak area in newly produced batches.

The validation was more focused on using the method as a limit test for aggregates, rather than calculating the precision at different concentrations. It was acceptable, but not fantastic. The column is a TSK column.
24 posts Page 2 of 2

Who is online

In total there are 30 users online :: 1 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry