PHOBIUS » Mon Sep 01, 2008 8:36 am
Dear Lusby,
I have recorded similar problems.
The mobile phase I am using is basic (20 mM acetate ammonium pH=10), therefore decreasing a silica based column´s lifetime rapidly.
I won´t write you much about the theory (silica dissolving, hybrid sorbents etc.) I rather share my practical experiences. I have tried different RP columns.
Synergi fusion (max pH 10) I used 9.2 Number of injections till peak broadening = 150
Gemini NX (max pH 12) I used 10.0 Number of injections 100 (!)
Xbridge (max pH 12) I use 10.0 I have done 100 injections and the column is still working perfectly. (will the fun ever stop?

)
I am not sure, if it helps you. The method is working, however it will be probably "column consuming".