						It is possible to calc EP s/n in the truer sense using Empower (from the blank injection), we just didn't have much success in setting up the custom field. If you can request Waters technical document #TECN1852625 it has how to do it and gives better explanation of how the algorithm works.
EDIT: The only concern I have is from the relationship S/N = 1/RSD. It's not an exact expression for HPLC, but I do use it as justification of specs and results. I would not in any way advocate the USP calc, which is quite vague and open to interpretation (why RSD is the fundamental basis for accepting or rejecting in the validations we run here). The EP calc isn't really noise, but a range of noise as can be seen in the definition of h. I would be quite surprised if anyone has calculated noise for a USP product using 20*w(1/2) as their range (unless it's a VERY sharp peak). Usually, at least in data I've reviewed, people tend to ignore single excursions and attempt to focus on the bulk of the noise. In the last EP verification study I ran it was like 5 minutes of baseline to monitor. It would seem to overestimate true noise IMHO, but is it 2X? I have the report in front of me, and I had a s/n of 13 with a 13% RSD (s/b around 5% RSD). Based on the single point, it would seem that they really are doubling true noise, but it could have been a number of things (including carry over, which the Waters equipment is notorious for).