Page 2 of 2
Re: If you need help with Waters HPLC/UPLC instruments, ENTER
Posted: Tue Apr 30, 2019 11:04 am
by SHERIFV4
Hello
https://www.waters.com/webassets/cms/su ... 035699.pdf
you can contact your Waters agent office in your country that you was purchase this column from him for certificate of analysis, and he will send you for this certificate
Re: If you need help with Waters HPLC/UPLC instruments, ENTER
Posted: Sun Jun 09, 2019 3:52 am
by andrewmax123
Hi
If you need any help for any Technical problem related to Waters HPLC , UPLC, MS systems, Feel free to contact me..
I provide full maintenance/ Qualification services/Free telephone support
I'm Waters service Engineer (with +8 years experience): "Chromatoghraphy@gmx.com"
Hi,
I have a question. what happens when you do not remove stainless steel sinkers when washing a UPLC system with 30% phosphoric acid? Does the phosphoric acid hurt the sinkers? I know that HASTELLOY SINKERS must be removed.
Thank you
Re: If you need help with Waters HPLC/UPLC instruments, ENTER
Posted: Fri Jul 12, 2019 7:26 pm
by SHERIFV4
Hello
In case of using phosphorice acid for system acidic cleaning :
You should remove the solvent filters, and remove any peek fitting or peek tubing (if used in your UPLC system), in addition to you need to disconnect the tube to flow cell of TUV/PDA, which mean you should not pass any phosphorice acid to your flow cell of the detector
This will increase life time of flow cell, any also will avoid any distortion in any peek part in your system..
Regards
Re: If you need help with Waters HPLC/UPLC instruments, ENTER
Posted: Thu Sep 12, 2019 2:38 pm
by MikeL42
My lab inherited a Waters 1515 Isocratic pump. I have been working with it occasionally the past couple of years. Recently it appears that the pressure transducer has become clogged. Mobile phase is not coming out of the transducer.
How difficult is it to replace a transducer? Any chance that I can fix this problem short of replacing the transducer? This pump is not under a service contract; it sees little use.
Re: If you need help with Waters HPLC/UPLC instruments, ENTER
Posted: Tue Feb 18, 2020 11:17 am
by fareedtayyeb
Can give a small push we are near but cant get the Console display
We are experiencing partial display on UPLC Control console. We are using Empower2 2154 with installed 1.65 ICS. MS FrameWorkNet 3.5v. It shows NOT CONNECTED with Red dot on both BSM,ASM also configured by DHCP,MAC address etc. Surprising thing is that our instrument activates Pump and autosampler works still showing not connected sign. WinXp with SP 3 .3.5FrameworkNet is also installed. The moment v select the Run Sample it appears with Black Bottons flashing but in fraction of seconds and then Red light and Not Connected appears. However v manages to run flow rate as per program and acquisition window also displays that its in operation. But v cant stop the system neither the Sampler Green button do anything. Can you guide us from this now.
Plz guide from here. Its our first time with Acquity system
Thanks and Best regards
Hi
If you need any help for any Technical problem related to Waters HPLC , UPLC, MS systems, Feel free to contact me..
I provide full maintenance/ Qualification services/Free telephone support
I'm Waters service Engineer (with +8 years experience): "Chromatoghraphy@gmx.com"
Re: If you need help with Waters HPLC/UPLC instruments, ENTER
Posted: Tue Feb 18, 2020 11:24 am
by fareedtayyeb
Can give a small push we are near but cant get the Console display
We are experiencing partial display on UPLC Control console. We are using Empower2 2154 with installed 1.65 ICS. MS FrameWorkNet 3.5v. It shows NOT CONNECTED with Red dot on both BSM,ASM also configured by DHCP,MAC address etc. Surprising thing is that our instrument activates Pump and autosampler works still showing not connected sign. WinXp with SP 3 .3.5FrameworkNet is also installed. The moment v select the Run Sample it appears with Black Bottons flashing but in fraction of seconds and then Red light and Not Connected appears. However v manages to run flow rate as per program and acquisition window also displays that its in operation. But v cant stop the system neither the Sampler Green button do anything. Can you guide us from this now.
Plz guide from here. Its our first time with Acquity system
Thanks and Best regards
hello,
Is there way to transfer HPLC method to UPLC method?
thanks
This has been covered here and at many other sites. Check the sticky postings, references and google for guidance. There are also online calculators that can help you.
Re: If you need help with Waters HPLC/UPLC instruments, ENTER
Posted: Tue Mar 03, 2020 6:59 pm
by fzawadiak
Hi, I need to replace tubing in Column Manager (old one) as it got plugged for good... How on earth can one access this low pressure valve? I spent two hours today trying to figure out how to disassemble it, would be easy if somebody thought and put screws/nuts under valves other way around. As it is I see no way to unscrew it, as the middle (cooler) part interlocks with valves. Any help?
Re: If you need help with Waters HPLC/UPLC instruments, ENTER
Posted: Tue Oct 22, 2024 4:44 pm
by bluedevils1999
Hi,
Our XevoTQD hooked up to Acquity H-Class UPLC is not communicaitng with Masslynx. We can't conduct EPC reset through the hole since that button is not responding (not making any beep noise). We can't do EPC reset through command lines since it is not pinging EPC (MS)...
Since soft-reboot wasn't possible we tried power cycle (venting and recycle) but it is still not connecting to masslynx or showing up on DHCP server.
Please let me know what I can do. Thanks
Re: If you need help with Waters HPLC/UPLC instruments, ENTER
Posted: Wed Oct 30, 2024 8:47 pm
by THNDRacket
Yes, I really hate how the reset switch is designed on the Xevos.
How about through Console | System | Troubleshoot | Scan Instruments
Does it give IP addresses for the MS and all modules? Perhaps a cable is loose or the IP address was corrupted somehow?
I believe I had a similar problem in the past. Not only did we have to power cycle, but also unplug the MS power cord for 5-10 minutes to give time for the capacitors to discharge.
And for others who might be stumbling across this topic, there are several ways to reboot the epc. Through the pinhole on the front of the mass spec; through Tune | Setup |Communications Status; and through Hyperterminal (or equivalent) reboot command.
Much of this may apply as well:
https://support.waters.com/KB_Inst/Mass ... n_MassLynx
Re: If you need help with Waters HPLC/UPLC instruments, ENTER
Posted: Wed Oct 30, 2024 8:50 pm
by THNDRacket
Yes, I really hate how the reset switch is designed on the Xevos.
How about through Console | System | Troubleshoot | Scan Instruments
Does it give IP addresses for the MS and all modules? Perhaps a cable is loose or the IP address was corrupted somehow?
I believe I had a similar problem in the past. Not only did we have to power cycle, but also unplug the MS power cord for 5-10 minutes to give time for the capacitors to discharge.
And for others who might be stumbling across this topic, there are several ways to reboot the epc. Through the pinhole on the front of the mass spec; through Tune | Setup |Communications Status; and through Hyperterminal (or equivalent) reboot command.
Much of this may apply as well:
https://support.waters.com/KB_Inst/Mass ... n_MassLynx
Hi,
Our XevoTQD hooked up to Acquity H-Class UPLC is not communicaitng with Masslynx. We can't conduct EPC reset through the hole since that button is not responding (not making any beep noise). We can't do EPC reset through command lines since it is not pinging EPC (MS)...
Since soft-reboot wasn't possible we tried power cycle (venting and recycle) but it is still not connecting to masslynx or showing up on DHCP server.
Please let me know what I can do. Thanks
Waters 2998 PDA Det Wavelength Calib Error
Posted: Sun Jan 12, 2025 10:50 am
by fareedtayyeb
Well I have replaced new lamp and new flow cell but 2998 never eve passes wavelength Calibration test neither the Erbium wavelength values gets closer to required wavelength ranges as mentioned in OP guide..
Flow cell is firmly slot in with guide pin also D2 lamp is properly installed.
On startup green lights comes normally then after few moments and erbium.filter motor movement it appears with red light with D2 lamp still glowing and give Wavelength Calibration Error
Tried almost all remidys changing solvents from water to formic acid,IPA,ACN etc.
Re installed the Firmware and ICS
I WISH TO KNOW ON STARTUP WITHOUT ANY ETHERNET CONNECTIONS WHY THE DETECTOR GIVE THESE ISSUES ONCE WE HAVE NEW FLOW CELL AND LAMP.
ALSO WHAT SHOULD BE THE READ ENERGY TYPICAL VALUES ON ERBIUM AND CLOSE POSITIONS IN 2998 CANSOLE AT 254NM OR 190-500 OR 190-800 NM SETTINGS ITS NOT MENTION ANYWHERE IN 2998 OP OR ANYOTHER LITERATURE AVAILABLE FROM WATERS
PLZ GUIDE WE HAVE COUPLE OF UNITS SITTING SICK WITH SAME PROBLEMS
REGARDZ