Ok, so....
I've done the "to test the filaments":
4) Measure the resistance between FIL A and FIL test points. meter says:
OL(out of range) no movement of "needle". i guess this means that it's less than one?
(manual say it should be less than 1 ohm)
5) Measure the resistance between FIL B and FIL test points. meter says:
0.3ohm
(manual says it should be less than 1 ohm)
6) Measure the resistance between FIL A, FIL B, or FIL test point and GROUND. meter says:
OL(out of range) no movement of "needle" on all.
(manual says: resistance should be greater than 1 megohm, lower resistance indicates one of the filaments is shorted to ground)

Measure the DC voltage between the FIL test point and GROUND. meter says:
-0.255V dc/107 k ohm
(manual says: impedance of your meter must be greater than 1 megohm, voltage should be about -70Vdc, if this voltage is not present then replace the filament cable.)
9) Measure the AC voltage between the FIL test point and the FIL B and FIL A test points. meter says: 0.006V for both.
(manual says: should be about 1V but any ac voltage can be regarded as a successful test)
Also Ive checked the MSD electronics as follows:
1) set electron multiplier to 0.
2) turn MS electronics on (MSON).
3) read MS status: says "
There is no emission current"
4) set electron multiplier to above 1000V (1400V) if the fault returns the problem is probably in the Top Board or in the wiring of the vacuum manifold. fault is still present!
So what would you start with?
Top board replacement? ($1450)
or replace filament cable?
or replace filaments?