Posted: Thu Oct 21, 2004 7:52 pm
HWM,
Wayne seems to have followed everything you suggested but the problem still seems to be there. Furthermore, in my mind, the likelihood of a gas/slow outgassing problems decreased from the moment he used 100% water as mobile phase and at the same time degassed his mobile phase (100% water have much less degassing problems than a water/ACN mobile phase...).
Einar seems to indicate that you can have sudden drops as well in addition to the drifting if you have lump or electrical problems with your UV. A change of the UV would be a straightforward way to know if the detector is to blame or not. Also I asked about the UV length as UV lamps at the end of their life are more trouble to use at low wavelengths (i.e. 190 nm) than higher wavelengths.
What Chris suggested is also a possibility and the quick fix is to use a longer or thinner diameter tube at the outlet of the UV cell (that goes to the waste). Another way is to buy commercial calibrated backpressure columns (from upchurch for example) of 50 psi or equivalent. However, among all the cases that I have seen this problem, the side effects were spiked peaks in the UV chromatogram... but again I do not eliminate completly the possibility...
Wayne seems to have followed everything you suggested but the problem still seems to be there. Furthermore, in my mind, the likelihood of a gas/slow outgassing problems decreased from the moment he used 100% water as mobile phase and at the same time degassed his mobile phase (100% water have much less degassing problems than a water/ACN mobile phase...).
Einar seems to indicate that you can have sudden drops as well in addition to the drifting if you have lump or electrical problems with your UV. A change of the UV would be a straightforward way to know if the detector is to blame or not. Also I asked about the UV length as UV lamps at the end of their life are more trouble to use at low wavelengths (i.e. 190 nm) than higher wavelengths.
What Chris suggested is also a possibility and the quick fix is to use a longer or thinner diameter tube at the outlet of the UV cell (that goes to the waste). Another way is to buy commercial calibrated backpressure columns (from upchurch for example) of 50 psi or equivalent. However, among all the cases that I have seen this problem, the side effects were spiked peaks in the UV chromatogram... but again I do not eliminate completly the possibility...