Re: quantification using a calibration curve
Posted: Fri Jan 19, 2018 10:57 pm
Thank you very much for the information.
Chromatography Forum is a public discussion group where you can post questions, news, or messages of interest to chromatographers everywhere, but you must be registered to participate (registration is FREE). If you are a registered user, please log in.
http://www.chromforum.org/
Bravo! I will never understand clients so wrapped up in LOQ's and LOD's when all the samples have ppm levels in them. When you are pushing the sensitivity of the instrument they may very well be appropriate but I'm not shooting a 50 ppt standard in a study where the low level is 10 ppm just to give someone a low LOQ. It makes no sense and jeopardizes the success of the study.What I have found for best peak shapes is to have the dilution made in the beginning mobile phase, this makes sure there is no mixing problems with the injected sample and the mobile phase that is entering the column at that time.
The more aqueous the better normally. If you have a beginning mobile phase of 90% aqueous and inject a sample that is 75% methanol, some of the analytes will ride that solvent slug through the column and give a large peak early in the chromatogram, then the portion that is retained will give another peak in the gradient. This is known as "Peak Surfing" as a portion of the analyte rides through the stationary phase like a surfer on a wave.
If you always do a dilution on the sample, then the LOD and LOQ will have to take into account that dilution amount, so there is no need to calibrate to the lowest possible amount that can be detected on the instrument if the sample will always have a high level detect. If the sample will always be 100ppm then you don't need to calibrate to 0.000001ppm just because the instrument can see it(for an example). You could easily calibrate as 10ppm, 50ppm, 100ppm, 200ppm, and 500ppm to satisfy the needs of the test. If there could be some samples that will be non-detect even when not diluted then you need to know the limits of the instrument and procedure.