Hi Adam

The problem with keeping the flame lit with 450 ml/min of air suggests something not right with the flows in the detector. Check the following: no obstructions in the flame tip and that the end of the column is not actually through the hole in the tip, that the flame tip is properly seated in the injector base, and that the gas supplies are connected to the right pipes - you might be putting air through the jet or something.

Peter
Peter Apps

Hi All,

Many thanks for all your help and suggestions. Hopefully I am on the way to getting this fixed.
I reinstalled the column and fitted a new fid jet. I now can get the fid detector to stay lit at an air flow of 400 ml/sec.

I now get a response of 10s of thousands of millivolts to the solvent peak, and my peak shape no longer tails.

Cheers
Adam
17 posts Page 2 of 2

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry