The detector should turn the suppressor off when receiving a TTL signal at the appropriate input. For some reason, this detector would not do that. Nor would our other Dionex detector. I could use the TTL signal to change the detector range or trigger a recorder mark, but the suppressor turn off feature would not work. It seems rather arbitrary that this feature alone would not work on multiple detectors.
So I set up two methods on the detector: one to run the analyses with the suppressor at the appropriate setting, the second method with the suppressor off. The detectors will increment method numbers using a TTL signal, so I finally got it to work this way.
Thanks for everyone's help and input.