 - Posts: 276
 - Joined: Tue Sep 07, 2004 10:05 pm
 
1: You can not use non-volitile buffers (ie no phosphate)
2: Non liniear stanadr curves (standard curves fit to a quadratic equation), not really a limitation, just the way it is
3: Limited sensitivity for volitile and semi-volitile compounds
4: Sensitivity negativly impacted be particulate in mobile phases and column bleed
5: Detector must be vented
6: Need a gas supply (most use nitrogen, some can run qiute nicely off of clean dry house air)
7: Destructive technique
Of course on the other hand, these detectors can detect compounds that can not otherwise be detected easily. Many, many people use these detecor every day without issue. You seem to be having trouble deciding what brand to buy, arrange to send these companies some sample or arrange for an on site demo, most vendors will be happy (mostly happy) to comply.
And finally to answer you question "Can this Sedex ELSD detector work for mobile phase with some aq. buffer or water, e.g. 55% buffer? "
If the buffer is volitile and the flow rate is within the range of the detector (most operate quite nicely at usual HPLC flow rates), then the answer is yes.