Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
It would be true if the response was measured in peak height.Using methanol one usually gets longer retention times than using acetonitrile in the same percentage. Longer retention times usually give broader peaks. UV detector is concentration dependent and broader peaks will give you higher area.
Sorry! I meant peak breadth/width and not height.It would be true if the response was measured in peak height.
In short, MeCN and MeOH will shift your aromatic hydrocarbon (what is the func. group here?) differently with different max abs. wavelengths, and will give you different detector outputs.Solvents can interact strongly with certain solutes and thus change the observed UV-VIS spectra, either by removing vibrational fine structures, or by shifting absorption band maxima, or both. When the UV spectra of phenol in water (...) and in isooctane (...) are compared, both a blue shift and the disappearance of splitting can be observed in the water case. The blue shift is typical for the stabilization of the ground state by hydrogen bonding solvents such as water (why?) , the splitting can only be observed in water-free solvents without hydrogen bonding (why?). Red shifts of band maxima can be observed in polar solvents when the excited states are more polar than the ground states...
,Note that column was not allowed to condition more than a couple of minutes
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.