Posted: Fri Oct 26, 2007 8:02 am
The aim of selecting the reference wavelength is to correct the signal obtained with the measured wavelength. Thus the reference wavelength chosen should be a wavelength where no one from the analytes analysed / mobile phase constituents absorbs on it.
So if you choose 210nm or 215nm as reference wavelength where background absorption is significantly high (especially when using methanol as modifier in the mobile phase), you'll get less sensitivity and also baseline that is prone to changes as mobile phase ratio changes.
Also using 280nm as reference wavelength when your analytes have absorption maxima near 280nm, will cause similar changes regarding sensitivity or noise.
Hope this helps