lmb » Fri Aug 17, 2007 3:38 am
Dear Peter,
I agree that there are two issues in our discussion.
One is how one should choose a column pneumatic conditions (pressure, flow, velocity, etc. of a carrier gas) when one begins to develop a new method, and another is how to maintain robustness of and existing method.
My previous posting addressed only the first issue, suggesting that the starting point should be selection of gas flow rate rather than its linear velocity. The flow rate could be selected according to formula (*) in my previous posting.
I would like to add the following to your comments on maintaining method integrity.
Of course, selection of the flow rate is not the end of the story. It is a good idea to compare expected and actual values of pressure and hold-up time. This might help to verify integrity of the system and to establish the control points for the system maintenance.
An important control parameter for verification of system integrity is hold-up time. For the system integrity, hold-up time is far more important than, for example, gas flow rate or linear velocity. For example, in some applications, a column is trimmed from time to time. To have the same retention times of all peaks after the trimming as they were before the trimming, one only needs to restore the hold-up time by reducing the column pressure. It is important to stress here that restoration of average velocity of a carrier gas would not restore the retention times.
You probably noticed that here I go again agitating against practical usefulness of average velocity. To set up pneumatic conditions of a carrier gas, use its flow rate (in mL/min). Once the method development is complete and system integrity is verified, measure hold-up time and keep it constant. It can be concluded that in both cases – method development and its maintenance – there is always something better than carrier average velocity. I really believe that it would be very healthy for practice of GC method development to pretend that the concept of average velocity does not exist.
[quote]“does this make separations run at these flows [50, 30 and 15 cm/s] more robust to column deterioration and poor inlet performanceâ€