I also have the "pleasure" of working with Beckman equipment. I have asked them numerous times about degassing and their only reply is: they have a special pump design so you don't need to degass your buffers! (amazing how a they can declare the laws of physics Null and Void)
What I do frequently, whenever I suspect bubbles in the detector, is place a 1 or 2 ml syringe on the outlet of the detector. Run the system and monitor the baseline. Allow the syringe to fill with mobile phase, then loosen the inlet to the detector (at the column) and force the flow backwards with the syringe. (I collect the few drops in a paper towel) Then tighten the connection again and observe what happens to the baseline.
I always vacuum filter my buffers, then sonicate with vacuum for ~ 10 minutes. I find that I can only run the system for about 2 days before I start to get funny baseline or pressure pulsing.
This is only a small piece of the "pleasure" I have with my current HPLC system - I can not express my frustration in language suitable for a public forum!