One other thought: There is really a variable mismatch of sample solvent/matrix and mobile phase.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Maybe - usually even - but possibly not. I can see a scenario whereby a significant portion (but not all) of the crud is washed off with his gradient and that successive sample injections yield what he's seeing because a good sized slug of crud is hitting the column with each shot and building up, but is not very tightly bound to the column. Running a blank after some sample injections could conceivably clean up the column enough to yield reasonably reliable results for subsequent standards. This might at least be consistent with his second peak being not nearly so affected as the first. It might be instructive to have a look at what's going on at a lower wavelength.I am not sure, anymore, of what is described here. You said somewhere that a standard injection is normal even after abnormal sample injections. That would rule out that a builtup of crud causes the rt shifts and peak broadening indicated in the chromatograms. .