Although the split ratio is set on the GC and is set at the same value for each injection, the actual split ratio that you get with each injection depends on the gas flow into the column and the gas flow out of the split while there is sample in the inlet. The flow and pressure controllers on the GC inlet have a limited ability to maintain pressure and flows at the set values when the flow and pressure suddenly increase and decrease as you inject the sample. Disruptions are minimised by smaller sample volumes (which give smaller peaks) and slower injections (which give wider peaks).
In principal an autinjector will give better repeatability, because it gives consistent injection speeds, but off hand I cannot think of one that will be able to take the sample from the 1 l gas bulb, although I suppose you could rig a CombiPal, MPS2 clone type by setting the syringe positions specifically for this job.
An alternative to syringe injections would be to use a 6-port valve, but you are still not going to be able to get down to the necessary detection limits in real samples.
The problem boils down to this; at the concentration levels that are relevant, the volume of sample that contains enough analyte for the detector to consistently measure is larger than you can put onto a capillary column without causing volume overloading (broad peaks). Therefore you have to reduce the volume of the sample without reducing the quantity of the analytes. You cannot do this by cold trapping because real samples contain water vapour, and so sorbent traps are the only practical way of getting the job done. Yes they have drawbacks, but these are more tractable than the problems with alternative approaches - some of which you have already experienced.
Good luck
Peter