To determine if I have a reasonable method, the first thing is to count the number of peaks. However, I don't know how many peaks to expect, have no idea what compounds they are either. If I am lucky, I can get the synthetic starting materials and intermediates to check the separation of known compounds. I don't use MS during the initial column screening. I do use the UV peak purity test but it is not always right. the UV spectra of the structurally related impurities are often very similar. peak area is an important parameter in tracking those peaks. peak shape also gives some indication if there's any coelution. since I use different columns and different mobile phases (each at two different gradient times), I will compare chromatograms from different conditions and try to decide which peak correlate to which. It gets confusing sometimes, especially when the elution order of some impurities witched under different conditions. Experience is very important here.
when do I stop? first stop is when the samples start to pour in. I will use whatever condition I believe is the best to get the results out. based on what I see in these new samples, I may have to go back to modify my method to separate additional impurities. the synthetic process is also evolving during this time. LC-MS is often used to determine impurity structures so the chemists can understand the reacions better. after seeing samples from several batches, I have a good idea which synthetic impurities are here to stay and which one can be removed during purification. I will finalize the method with the most representative sample (separating the most number of peaks). This takes weeks to months, and a good working relationship with the chemists is critical.
Once the synthetic impurities are taken care of, it's time to check out the degradants through "forced degradation" studies. again I don't know how many peaks to expect. at this time I usually do it with LC-UV-MS. peak purity is a major parameter, especially for the main peak. Watch for any new peaks in the chromatogram, as well as any significant peak area change in existing peaks from the unstressed sample. If there's any indication of a degradant coeluting with a known peak, I have to modify the method again.
finally, the most labor intensive part: method validation. the specificity is pretty much proven during method development. we just have to make many solutions to check out linearity/range, sensitivity, precision, accuracy (spike recovery), solution stability, robustness, etc. then try it with different columns, instruments, analysts, and for different days. Appropriate system suitability parameters will be determined. The method is written into an official document. Now we are ready to test the clinical materials, and then the stability samples.
I hope I’ve provided sufficient justification for my paycheck.