Choice of mobile phase

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

18 posts Page 2 of 2

Thanks to all for your comments. I am not testing a protein - it is a new pharmaceutical API and as such I cannot say too much about the compound. We've switched to a 250 mm, 5 um Luna C8(2) column using a mobile phase of 50 % ACN:50 % pH 5.5 Buffer. The chromatography is fairly good (symmetry and USP tailing both 1.0, high plate count). We're just looking at shortening the run time.

Great stuff Mojo, don't you love it when a method comes together?!?

I know I do... 'see that? That resolution you could drive a bus through? I did that! :-D'

When that doesn't make me happy any more it will be time to find a new job.

Mark, Thanks
Excel