Hi all,
Thanks for the help so far.
I will post up progress and summary so far of the method:
Stated Method details
Column Details: SPB-1 Fused Silica Capillary Column (30m x 0.25mm x 0.25um)
New column and cannot cahnge it
Oven Temp = 60C Isothermally
Spilt Ratio = 1:50
Progress so far
GC Make & Model: Schimadzu GC-14A - Manual injection
Column Details = as above
Column Make = Supelco
Syringe Size = Glass 1ul
Sample size injected = 0.1ul
chemical components =
Butan-1-ol, 116 -118C boiling point (bp), Retention time (RT) = 3.78min, Area = 348877
IBMK (internal Standards), 117-118C bp, RT = 4.94mins, Area = 349099
N-butyl acetate, 126-127C, RT = 7.66mins, Area = 317879
Best GC conditions so far
Oven Temperature (Temp) = 45C Isothermally
Injection Temp = 135C
Detector Temp = 170C
Carrier Gas = Helium
GC Flame (FID, flame ionisation detector) = Air & hydrogen
Packed column on detector & injector 2. So passing nitrogen down this column
Helium Gas Flow = 22cm/sec
Split flow = 2.63mls/ min
Aim: Studying the ratio of the areas of the butanol and n-butyl acetate peaks during a reaction.
Validating the method at the moment to obtain the best possible resolution and tailing for the three peaks. At the moment n-butanol and IBMK are not seperating out correctly. There is a trought between the two peaks however it is quiet close to the baseline. There is tailing on all the peaks espically the n-butyl acetate peak.
Butane experiment
thanks a mill Ralph. This trick is really good and fast.
Injected 5ul and 10ul of butane gas onto the column and the following results were achieved:
5ul Butane: RT = 2.306mins, Area = 168947
10ul Butane: RT = 2.246mins; Area = 340904
Using the above
Velocity of carrier gas =
22cm/ sec or 2.246min/30meters, 1sec/ 0.22m, 0.22m/sec Think this is right
Gas Flow (Soap Film bubble

)
Carried out this test. Thanks Bruce for explaining this.
Flow from split flow is 3.8sec/ 10mls
Which is 2.63mls/ min
The glass tube was graduated to 10mls.
I also noticed that there was two vents side by side. One vents states purge and the other vent states split. So I measured the flow from the split vent. There was no flow exiting the purge vent.
So should I alter any of the above parameters or should I just go with the peak shapes that I have.
Thanks a mill for the all the help that I have got so far on the forum. I really apreciate the help.
Michelle