The mixing of THF/Water creates a warm mixture, and I often see a slightly turbid solution. Hence my preference to premix and filter aqueous mixtures of solvents like methanol and THF. I wondered whether sample and mobile phase temperature changes were contributing to the insoluble material and split peaks.One of the earlier methods is water acetonitrile gradient, the other is water THF isocratic. The first one works well, the second one sometimes gives the same problem; not so quickly as the new method; sometimes it is possible to perform a complete analysis (about one day) without splitting the peaks, sometimes the column does not resist.
Multiple blank injections between sample injections did not help.
I still have to try injectind more diluted sample, rising the temperature (although the column manufacturer told that with higher temperatures the permitted THF quantity decreases) and to filter the mobile phase one hour after the preparation.
I will get back as soon as I have the answers.
I have not tried either using buffer solution instead of water. Since the compound is not ionic, I developed the method without buffers. However, it could be another thing to try.
Other questions I have relate to needle wash solvent ( do you use one, if so, what? ), and do you use a pump seal wash, such as IPA/Water?
The reason I suggested using a buffer ( besides desperation

Good luck,
Bruce Hamilton