Posted: Tue Sep 26, 2006 4:15 am
I think that is only correct if you have the same number of acidic and basic sites. In Molly's case it might be not true. If you have more basic sites than acidic you can consider compound as basic. If you have more acidic sites you can consider it as a acidic compound. Also if charges are separated by space you can introduce a stronger acid or base on the surface to interact with zwitter-ion. You also need to look at pKa values of your amines and acids.
In addition to that in 100% water you cannot adjust retention by decreasing amount of ACN because negative concentration of ACN does not exist. In case of mixed mode (Primesep)-you still can use ACN to adjust retention plus you have buffer concentration which will affect your selectivity. Why use RP mechanism if is not working in the first place. On the best RP column (Atlantis T3 as you say) you will have for example retention of 3 minutes vs. retention of 2.1 minutes but what’s the difference? If you have other compounds you don't have to much room (no room at all) to adjust selectivity/resolution.
In addition to that in 100% water you cannot adjust retention by decreasing amount of ACN because negative concentration of ACN does not exist. In case of mixed mode (Primesep)-you still can use ACN to adjust retention plus you have buffer concentration which will affect your selectivity. Why use RP mechanism if is not working in the first place. On the best RP column (Atlantis T3 as you say) you will have for example retention of 3 minutes vs. retention of 2.1 minutes but what’s the difference? If you have other compounds you don't have to much room (no room at all) to adjust selectivity/resolution.