cene » Tue Apr 24, 2018 1:37 am
													
 
						
					 
					
						I just realised that I forgot to follow up on this.
Dear,
Please do following.
1. Check inlet temperature it should be around 300 C
2. Check appropriate septum purge flow should be around 3 ml/min
3. Appropriate column gas flow 
4. Increase split ratio or decrease sample size ( Injection volume )
5. Appropriate makeup flow around 30 to 40 ml/min
I have set the inlet temperature to 300, septum purge flow to 3 ml/min, column gas flow to 1 ml/min, split ratio to 25:1, injection volume to 1 ul, makeup flow to 29 ml/min.
There was a suggestion earlier to increase inlet temperature to 300C. Did you try that ?
Trimming 3 cm will not help, take off about 30 cm.
Peter
Yes, inlet is at 300.
I am still getting tailing peaks - the same as in the previous picture I uploaded. I take it that I just need to trim more length off the head of the column?