Re: wine analyse with gc
Posted: Sat May 13, 2017 11:36 am
If you are working from an unknown product you will have to run an analysis to determine ethanol concentration then run the methanol test afterwards. Here we would just have ethanol and methanol in a standard and run them both at the same time in one injection.@ dblux_ The fact that spirit drinks are excisable goods has led to the fact that the values of concentrations of volatile components need to be determined in mg per liter of absolute ethanol [mg/l (AA)] (see REGULATION (EC) No 110/2008 and COMMISSION REGULATION (EC) No 2870/2000). A detailed description of this circumstance is presented here https://www.youtube.com/channel/UCXgL2c ... lW1oxOGqtQ .
@ James_Ball The concentration in mg/l can be easy obtained by the following simple formula:
Concentration of methanol[mg/l] = concentration of methanol[mg/l (AA)] * “strength [%]”/100[%].
As a consequence, in the case given by You with approximately 15% ethanol, the calculated value of the methanol concentration in the dimension of the [mg/l] will be 15/100 times the concentration in the dimension of ”[mg/l (AA)].
Regards,
Siarhei
Dear James_Ball,
carefully follow the text.
The volume content of ethanol in the unknown alcohol product is confidently known and is equal to 789300 mg per liter of ethanol. As a consequence, it is possible to determine the methanol content of an unknown sample in one measurement on a gas chromatograph with equipped with FID.
Does not matter the type of drink and its strength. The possibility of a new proposed method makes it possible to calculate the content of volatile compounds in the sample quite accurately at once http://inp.bsu.by/calculator/vcalc.html .
The competition of GC manufacturers has led to the fact that the values of the RRF already are very stable for each model of the GC and they could be tabulated.
The procedure for introducing the IS substance into the analyzed sample is completely absent. And after that is it possible to give up such happiness?
Regards,
Siarhei