Leonardo, you made a general claim based on boiling point, i.e. not restricted to the specific samples I am running, so what difference to your general claim do my samples make ?Peter,
I write "it could happen" because I don't know the nature of your sample and method you are using. To determine how it will affect the retention time of your sample is necessary.
In my experience with samples that I used to analyze, the gain in separation using a ramped temperature inlet fully compensates a shift in retention times.
And how thohry said, the method that he follows requires an on-column inlet, so it was proven that for his kind of sample an on-column inlet works better than a S/SL inlet. Probably because an isothermal temperature in inlet can break down some molecules in sample.
Leonardo.
Do you have any hard evidence (i.e. not your experience) that chromatography is improved by delayed transfer of heavy compounds from a GC inlet into the column ?, or any hard evidence of retarded elutions under such conditions ?
I am harping on about this because separations are what chromatography is about, and your claim of improved separations contradicts accepted practice, which is to start the separation on the column with all the analytes in a narrow band near its inlet end.
Peter