Re: GC/FID recovery dropped
Posted: Tue Jun 13, 2017 5:48 am
"I want to believe that my inlet is leak free because the small peaks I actually am seeing are very sharp with minimal tailing or fronting."
A leak is by far the most likely explanation. Check the system with a leak seeker. Sharp peaks do not mean no leak - a splitter is a controlled leak and it gives nice sharp peaks.
Peter
A leak is by far the most likely explanation. Check the system with a leak seeker. Sharp peaks do not mean no leak - a splitter is a controlled leak and it gives nice sharp peaks.