YM,
It seems the difference is setting of PDA detector and UV. Could you pls check , if you are using any reference wevlength in your PDA?? if yes turn it off and rerun your samples. I have seen if your peak of interest fall between reference wevlength range it will not show in PDA, whereas in UV there is no such setting.
JM
It seems the difference is setting of PDA detector and UV. Could you pls check , if you are using any reference wevlength in your PDA?? if yes turn it off and rerun your samples. I have seen if your peak of interest fall between reference wevlength range it will not show in PDA, whereas in UV there is no such setting.