bubble on top of column

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

24 posts Page 2 of 2

Generally, silica-based packings are fairly robust with respect to pressure (unlike resin-based packings). A void space and high pressure may have related causes, but high pressure by itself won't usually collapse the bed. Your 4000SW column, with its big pores can be an exception to that generalization.

It's easy enough to check. Dig out a couple of mm worth of packing from the column inlet, replace the end fitting, and test the column to see if the pressure drops back down. The column is dead already, so there's little risk of making things worse.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

hmmm...interesting idea...worth a try. i will try it sunday or monday.

so you are saying if i take out what i think is void that may be causing pressure problem and then run flow thru it and if P comes down then know that it was the void and replace void by adding more resin? Can you clarify? How will having empty space on top of column and running flow through effect column? Will resin expand into empty space?

Sue
sargum

Hi,

I went in and scooped out some resin and am replacing with new silica resin.

Question: some of what i scooped out had a very jello-like consistency. Is this normal? What could have caused it?

S
sargum

Still would like to know how you got at this 4000SW material. Gel or wax shouldn´t be there.

The top layer of gel might've been destroyed by chemicals under high pH. Fine particles could be generated on top of the column. When you reverse flush the column, the fine particles could block your new frit and cause high pressure. Have you tried normal flow direction after you had the high pressure problem?

I got to the resin by unscrewing the inlet fitting.
sargum

YX:
Interesting theory about pH destruciton of silica. I will check my PBS eluent pH for kicks. It should be close to 7. Do I need to worry about pH issues with methanol? I did try and run flow in the downflow direction and had encouraging flow at first (for a half hour) but then subsequently the pressure went really high and you know the story. Yesterday, I unscrewed the outleft fitting just to see if it also had jello-like consistency (like the stuff near inlet fitting) but instead, I found dry wax-like buildup. It was deep too. The column is totally shot. I am just trying to unravel the mystery. So, you think that the silica got destroyed?

I have been using FITC-labelled dextrans and nonlabelled dextrans on the column. No honey! Does pH of the sample matter? At 20ul, it shouldn't right if PBS is eluting it thru? I did have an old guard column, which people point out may be the culprit.

I would really be intersted in hearing your answers/comments on these issues. I am really confused and frustrated.

S
sargum

YX or other: is it possible that resin looks like wax b/c no flow was getting thru, causing resin to dry up? I would luv to hear all possibilities.
sargum

Maybe some basic books on a some topics?
24 posts Page 2 of 2

Who is online

In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry