M-KW » Mon Aug 08, 2016 4:25 pm
Thanks for the replies so far by the way, I really appreciate your help!
These extreme conditions were tested and I found out that there is a noticeable increase in the response of the late PAHs. So I tested 250-340°C injection temperature and I found out that for the highest injection temperatures I get the best results. Given that they state I can go up to 350°C I thought 320°C was a good intermediate between what the septa can and life span. (Thermogreen LB-2 Septa, preconditioned)
However it seems that I misjudged and 320°C is too high... I will go down to 280°C and live more stable but with a bit smaller response of the higher PAHs.
The pressure pulse increased (now that I go without I can see it again) the performance of the naphthaline peak, so I will start using it again since that hasnt been the problem. (instead of 25 psi I use 300kPa (43 psi))
The septa of the vials dont seem to be the problem since the contamination was found also without injection, even in about the same concentration.
I have to contact my service partner to get to know if there are graphite o-rings to decrease septum bleed. I actually do use some standard o-ring (since they are delivered by the manufacturer, I thought they are for all purpose)
I am now at 290°C transferline temp and 300°C EI Source temp (was 280°C EI)
I get sensitivity loss first on last PAHs and then the earlier are getting worse too. Since this happened in about 6-10 injections I was able to see this decay.
If you found your liners to be perfect for this analysis I want them too

, did you mean these: Sky 3.5 mm ID Uniliner Inlet Liner w/ Wool (Catalog # 23459.1)?