Helium or nitrogen as make-up gas for capillary GC-FID?
Discussions about GC and other "gas phase" separation techniques.
16 posts
Page 2 of 2
What sort of GC are you using that needs a make up gas to get FID to work ?
Standard capillary columns operate at about 2ml/min, micro columns for fast GC may be 0.1-0.5ml/min not a big flow difference when you are looking at 300ml/min of air and 30ml/min Hydrogen. A good FID design does not need make up gas
. Use of Nitrogen make up can be used to optimise the flame for the most sensitivity however the difference is not great. I thnk that at one time at least one gas supplier provided special Nitrogen enriched FID air here in the UK, I do not know if many used it and if it is still available.
16 posts
Page 2 of 2
Who is online
In total there are 61 users online :: 0 registered, 0 hidden and 61 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 61 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 61 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science