Ok, so I hadn't realized the Agilent tech changed the inlet type from splitless to split while he was here on Friday, and that the temps were all turned down. So, remedied that for a better comparison. All our methods are splitless injection.
So:
The older system is an Agilent 6890N/5973 GC/MS with EI source. It's gas line is piped from the helium outlet, then two scrubbers (a universal trap and a really old, unlabeled trap, maybe hydrocarbon), then to the EPC. The total flow on the older system is 20. Column flow is 2 ml/min. Inlet temp is 280, oven temp is 60, aux temp is 300. Source and quad are at 230 and 150. Nitrogen level is ~2.5% of the 69 ion right now.
New system is a 7890B/5977 with HES source. Gas is from the helium outlet next to the older system, T'd off, with no scrubbers. Total flow is 25. Column flow is 2 ml/min. Inlet temp 280, oven temp 60, aux temp 300, source and quad are 230 and 150. Nitrogen level is currently ~25% of the 69 ion.
Pictures:
So, from the helium tap to a T-connection. The left goes to the two scrubbers, the right goes to the new instrument.
They didn't have enough line to make it all the way to the EPC. So.. extension.
And then straight to the EPC.
Agilent was just here and found a very small leak in the T going to the new instrument (first pic). So he remedied that and now I'm giving it some time to purge out.