Posted: Sat Dec 31, 2005 5:48 am
I thank you all for your help and advice.
I'm now checking the flow rate during the day. The pressure is very stable.
As soon as I finish, I'll have two different scenarios:
1. the pump can't work properly at that fixed flow rate
or
2. the column isn't the best column, a naughty C8.
In any case I think I'll have to review the method.
A further info is that now the Rts seem to be steady and my standard peak is now set at 10.7 mins, more or less. When the column was new, this Rt was 12.5 mins.
It seems that the column took two weeks to equilibrate...
I'll let you know the results of the flow check, as soon as it is over.
'Pump can't work properly'...so this implies the flow rate must be unstable. This decreasing RT suggests volume out is more than expected. Collect from the waste 10 ml in flask. Set your stop watch and flow rate at 1.0 ml/min. Volume should be 10 ml if pumps are working properly.
See the valves of the pump...does piston wobbles while moving in or out.
If pumps are working properly, flow rate should be 1.0 ml when set at 1.0 ml/min.
Even an excess of 100 microliter at flow rate of 1.0 ml/min will shift the RT.
How can you say its naughty. Is it some technical term or what I am thinking naughty
Check the LCGC article over retention time changes.
http://www.lcgcmag.com/lcgc/data/articl ... rticle.pdf
Why will column take two weeks to equilibrate. Its juzt that you are not able to set a proper chromatographic conditions.
Flow rate, pressure, Mobile phase composition, temperature...
Regards,
Amaryl.