Re: Phenols reference standard dissolution
Posted: Thu May 12, 2016 9:28 am
HI Gugu
I have just re-read you post on how you calculate the volume of standard you are adding by weighing the syringe. You are doing the weighing wrong. The way that you are doing it includes the volume of standard in the needle of the syringe, and that volume is not added to the solution when you push the plunger down. When the plunger is pushed down the needle is still filled with standard - that volume of standard does not get into the solution.
That is (at least partly) why if you calculate the volume using the density of dichloromethane you will get a volume of 7.7 ul when the syringe reads 5ul - the difference in weight between the empty syringe and the syringe filled with standard and the plunger at 5 ul includes the weight of the standard in the needle. So your calculated volume includes the volume in the needle, which is not added to the solution.
You need to weigh the filled syringe with the septum on the needle, add the standard to the solution, recap the needle, then weigh again. The difference in weight is what was pushed out of the syringe into the solution when you pressed the plunger down to zero. The needle was filled with standard both before and after you transferred the standard to the solution and so that does not contribute to the difference in weight.
Try it the way I describe, calculate the volume of standard using the density of dichloromethane and tell us what volume you get.
Peter
I have just re-read you post on how you calculate the volume of standard you are adding by weighing the syringe. You are doing the weighing wrong. The way that you are doing it includes the volume of standard in the needle of the syringe, and that volume is not added to the solution when you push the plunger down. When the plunger is pushed down the needle is still filled with standard - that volume of standard does not get into the solution.
That is (at least partly) why if you calculate the volume using the density of dichloromethane you will get a volume of 7.7 ul when the syringe reads 5ul - the difference in weight between the empty syringe and the syringe filled with standard and the plunger at 5 ul includes the weight of the standard in the needle. So your calculated volume includes the volume in the needle, which is not added to the solution.
You need to weigh the filled syringe with the septum on the needle, add the standard to the solution, recap the needle, then weigh again. The difference in weight is what was pushed out of the syringe into the solution when you pressed the plunger down to zero. The needle was filled with standard both before and after you transferred the standard to the solution and so that does not contribute to the difference in weight.
Try it the way I describe, calculate the volume of standard using the density of dichloromethane and tell us what volume you get.
Peter