Endrin Breakdown, need help

Discussions about GC and other "gas phase" separation techniques.

17 posts Page 2 of 2
active site in your instrument

almost in inlet ( active site as result of high temp , when you lower temp of inlet you got something for this I think that active site in the inlet )
Are you running in split or splittless? If you are running splittless the sample has more time to come in contact with the inlet. This will promote breakdown of Endrin and DDT. Start at a higher oven temperature and use a high split flow. Using the CLP 1 & 2 from Restek, I start my oven at 120 C and ramp 9 C/min up to 300. Inlet temp is at 250 C. Split flow is 20:1 (I actually could bump it up a little if I wanted too). Running in split mode transfers the targets on to the column quicker and more efficiently and will reduce the amount of maintenance you will have to do. I can see a 20 ppb standard no problem.
17 posts Page 2 of 2

Who is online

In total there are 113 users online :: 1 registered, 0 hidden and 112 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 112 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry