I do not really know which ELSD you use, and there maybe differences due to the different nebulizers but:
When I worked with normal phase solvents (i.e. mixtures of hexane with other solvents/additives), I noticed things precipitating and immicibility problems in the nebulizer. Then I touched the nebulizer and saw that it was extremely cold. Actually what was happening (I think) is that evaporation of hexane is pretty fast and endothermous, which was decreasing the temperature of the whole nebulizer at such a point that you start seeing immicible phases. These immicibility issues, depending on your mobile phase, can increase your background noise or even the peak area of your analyte as during the analysis you will have less and less of your analyte reaching your detector.
In orded to test it, I heated the nitrogen lines that were used for the nebulization to about 50 C (that was the temperature of the nitrogen at the exit of the nebulizer) and no problems what so ever from that point.
I suggest that you follow this strategy (heating your nebulizer), first pass through 100% isopropanol (you do not have to use a column for this) for about 30 minutes, then you can use your normal phase solvent and your separations conditions. In addition, it might be a good idea to increase your drift tube temperature as well.
Once you do this, let us know if it worked (or not).