I figured out what the issue was today. Purge line was attached to split vent and vice versa during an earlier maintenance. Thus why the flow was setting right but not actually carrying away sample. Thanks for the help!
OK ! Thanks for the feedback - that's a new possibility to consider in trouble shooting.

Peter