The products we make are methacrylate monomers which contain an azo-based polymerization initiator. We need to determine how much initiator is present. A calibration has been set up using different concentration levels of the azo initiator, and the peak area represents a certain percentage. We don’t use an internal standard
Please give the details of how you did the calibration - what you dissolved in what, how you did dilutions, the range and the r-squared, and the repeatability if you have determined it.

What is the signal:noise of the peak that you are interested in ? Can you post a chromatogram - instructions are in a sticky at the top of the LC forum page.

Is the initiator stable at GC temperatures ? if you add an intrnal standard with a similar MW do you also get poor repeatability for its peak area ?

Do you have to use acetone as the solvent, or could you use something with a higher boiling point and higher MW ?; the pressure pulse that you get when acetone hits the inlet at 300C could be the cause of the poor repeatability.

Peter
Peter Apps
17 posts Page 2 of 2

Who is online

In total there are 170 users online :: 1 registered, 0 hidden and 169 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 169 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry