GC-MS air percentage increases with oven temperature

Discussions about GC and other "gas phase" separation techniques.

18 posts Page 2 of 2
...loosen the front end nut (start of column) and then make sure the reducing nut is tight (and should have a metal o-ring in it).
...
Just in case somebody is looking for a "metal o-ring" - he will never find it :lol:
However there must be a washer.
...loosen the front end nut (start of column) and then make sure the reducing nut is tight (and should have a metal o-ring in it).
...
Just in case somebody is looking for a "metal o-ring" - he will never find it :lol:
However there must be a washer.
yes, metal washer. forgot and used as close to a synonym as possible.
we are having some issues with 3 different LC/MS/MSs and that has been a little taxing so sometimes my vocabulary goes out the window.
thanks for correcting me.


KDM
Could be a cracked column. Try re installing the old one to see if your leak is fixed.
18 posts Page 2 of 2

Who is online

In total there are 80 users online :: 0 registered, 0 hidden and 80 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 80 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry