GC hp6890 peak response differences

Discussions about GC and other "gas phase" separation techniques.

17 posts Page 2 of 2
Hello

It is quite strange discussion...
You can not really compare 2 systems using absolute response for peaks. Below few thing that can affect absolute response:
- liner and glass wool position (evaporation problems for some compounds can cause different responses)
- makeup gas flow (small differences +/- 0.5 ml/min can affect peak height)
- electronics (you will never have 2 identical electrometers)

I agree with previous comments: you need to provide more information...and if you think that it doesn't matter my answer is: it matters :)

Best regards

Tomasz Kubowicz
Hi ,

Thanks for your reply.
this makes sense , though never had, from an HPLC point of view, an identical system that could not run a related substances method due to sensitivity.
This is the problem between the GC's (can only run one for res solvents)
17 posts Page 2 of 2

Who is online

In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry