from one the articles i have read, the ratio at 2 ml/min looks more 2-3 and around 1 for 0.5 ml/min (Chromatographyonline issue of May 2005, The use of acquity uplc in pharmaceutical development, figure 1 page 16), and the optimal flow range starts at 1ml/min.
my remark regarding column lenght limitation was on the base that out of the 3 parameters of column i.d., particle size and column lenght only the later would be changed. under those conditions there should be a maximum column lenght in theory.
the article actually "focuses more on another problem" (sorry, due to lack of a better term since it is not the articles main subject) of small i.d column which is their small internal volume and the effect to early eluting peaks.
this got me "trap" in contradiction:
the optimal range for the column according to the Van-Deemter plot starts at 1ml/min up, but the greater my flow rate then the earlier my peaks elute and then i have to deal with that broadning effect, then i have to find a second way to retain my peaks, which would be first of all to go and increase the buffer concentration. small i.d. of tubing, very small volumes of system, probably not a good idea in the long run for the system.
second option is too increase column lenght!
so know it looks to me that the smaller the column i.d. then columns must have a minimal optimal lenght and a maximum optimal lenght, in order to prevent peak broadning (eventhou the causes are different at both extremes).