Low-pH taling of bases in RPC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

18 posts Page 2 of 2
To confirm the silanol activity, I usually inject benzyltrimethyl ammonium chloride and see the peak shape for detecting silanol activity. If the peak shape is symmetrical and the elution time is not too high, it means there is little cation exchange going on. You can convince the reviewer by a similar experiment.
Thank you for the good suggestion. Actually, I found protonated histamine showed no retention and no notable peak tailing under the same conditions (see the figure below; sodium nitrate as void marker).
Image (The first peak represents histamine)
In my opinion, this may indicate that there was no significant silanol effects: If If any significant ion-exchange interactions had taken place, protonated histamine would have had some silanophilic retention on the column.
Any comments?
Thanks
As the name of your stationary phase implies, it has large steric groups (probably isopropyl) which protect the silica suface. Even if there are silanols, they must be shielded.
Thank you. A brochure from Agilent said:
“Agilent ZORBAX StableBond columns use patented, unique, nonfunctional silanes with bulky diisobutyl (SB-C18) or diisopropyl (SB-C8, SB-C3, SB-Phenyl, SB-CN, and SB-Aq) side chain groups that sterically protect the key siloxane bond to the silica surface from hydrolytic attack at low pH.”
To confirm the silanol activity, I usually inject benzyltrimethyl ammonium chloride and see the peak shape for detecting silanol activity. If the peak shape is symmetrical and the elution time is not too high, it means there is little cation exchange going on. You can convince the reviewer by a similar experiment.
Thank you for the good suggestion. Actually, I found protonated histamine showed no retention and no notable peak tailing under the same conditions (see the figure below; sodium nitrate as void marker).
Image (The first peak represents histamine)
In my opinion, this may indicate that there was no significant silanol effects: If If any significant ion-exchange interactions had taken place, protonated histamine would have had some silanophilic retention on the column.
Any comments?
Thanks
This experiment along with the concept of the bulky silane will be a good educational addition to your manuscript. Note that when bulky silanes are bonded, the surface coverage is usually lower, so you actually have more free silanols; it is the bulky group which acts like an umbrella to protect the analytes and the stationary phase being exposed to bare silica surface. It is not clear whether the manufacturer further end-caps after bonding with a small silane or not. By chromatographic characterization, it is pretty clear that the analyte cannot see silanols very well. The reviewer, like any scientist, must have made an educated guess, which in your specific case turned out to be incorrect. Good luck!
M. Farooq Wahab
mwahab@ualberta.ca
18 posts Page 2 of 2

Who is online

In total there are 22 users online :: 3 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Bing [Bot], Semrush [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry