GC-MS Method Development Question

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

21 posts Page 2 of 2
... Incase if I have to use carbondisulphide, how can I rinse carbondisulphide thoroughly between injections and what wash solvent can I use?

Please let me know.
Thank you,
Srikanth.S
It's a solvent. You don't need special procedures. Just use CS2 as wash solvent.
In general extracted ion chromatography has no benefit when analysing hydrocarbons as they tend to fragment in the same way (so all have similar spectra) and it is difficult to resolve isomers.

If you are looking at several individual isomers then quantitaion by an extracted ion is OK - run a calibration using standards (internal or external calibration)

If you are expecting hydrocarbon products then TIC is better and you should quantitate fractions of the chromatogram. the fractions represent hydrocarbons of different chain length and you can set the boundaries using n-alkane standards. You can either calibrate using a product of known concentration or use individual hydrocarbons and assume that the response factor applies to the whole fraction. External calibration is better for TIC. If you are adding internal standards you need to subtract them from the fraction that is being quantitated.
----suffers separation anxiety----
The method using CS2 will work as I ran it decades ago. If you have tha resources I feel it would be better to purchase a thermal desorber, then you could collect the samples on tubes packed with VOCARB 3000. The thermal desorber is similar to a purge and trap in that there is no solvent to worry about nor loss of volatiles when preparing with CS2.
Hi all, thank you very much for responses. We don't have a purge and trap capability. We have a split and splitless inlet with liquid autosampler. So, I cannot use thermal desorption procedure. Do any one have any idea how to semiquant quantitation using four or five standards?

I am thinking to go with semiquant procedure.

Please let me know.

Thank you,
Srikanth.S
I believe to semi-quant (quantitate multiple analytes from one standard) then you just set up your calibration as normal, but integrate the same peak for the different analytes in the calibration.
----suffers separation anxiety----
Hi thank you for the reply. Your suggestions are very helpful.

Srikanth.S
21 posts Page 2 of 2

Who is online

In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry