Re: Crude Oil Charactarisation
Posted: Mon Sep 01, 2014 8:37 pm
The problem working with this type of sample is that you have a lot of co-elutions. An alternative is calibrate the system with a n-Paraffins standard using a HP-1 column (a metal capillary column, because to bake the column from the C36+ you need a temperature higher than 400°C) using a SIMDIS system, but you will need to estimate the areas for each peak, because the baseline won't be flat in any part of the run.
Using a PONA column, probably a great part of your sample will remain in the column, even with the column bake and you will take a long time to finish the run.
For the light fraction, you will need to use the DHA technique, because the separation for light fraction (C10) is very poor in SIMDIS.
Using a PONA column, probably a great part of your sample will remain in the column, even with the column bake and you will take a long time to finish the run.
For the light fraction, you will need to use the DHA technique, because the separation for light fraction (C10) is very poor in SIMDIS.